Load Cell Drift in Scales

04/26/2010 8:47 AM

I have a precision scale that is used very seldom.

When it was calibrated ( annual) last February, several test points were OOT, too low ( by .0030g).

Question: Is this normal for load cells to drift this much?

In a CAPA, I recommended a certification ( taking it through its calibration routine) with each use.

Thanks for any input.

Register to Reply
Pathfinder Tags: CAPA Metrology scales
Interested in this topic? By joining CR4 you can "subscribe" to
this discussion and receive notification when new comments are added.

"Almost" Good Answers:

Check out these comments that don't yet have enough votes to be "official" good answers and, if you agree with them, vote them!
Anonymous Poster
#1

Re: Load Cell Drift in Scales

04/26/2010 9:19 AM

What is the precision and capacity of the scale?

Register to Reply
Guru
United Kingdom - Member - Indeterminate Engineering Fields - Control Engineering - New Member

Join Date: Jan 2007
Location: In the bothy, 7 chains down the line from Dodman's Lane level crossing, in the nation formerly known as Great Britain. Kettle's on.
Posts: 32175
Good Answers: 839
#2

Re: Load Cell Drift in Scales

04/26/2010 10:01 AM

Given that the test weights are slightly buoyant in air, would a change in air pressure since February, and hence the density of the air, and hence the buoyancy of the test weights in air, change the scale reading by this amount perhaps?

__________________
"Did you get my e-mail?" - "The biggest problem in communication is the illusion that it has taken place" - George Bernard Shaw, 1856
Register to Reply Score 1 for Good Answer
Guru

Join Date: Aug 2005
Location: Hemel Hempstead, UK
Posts: 5826
Good Answers: 322
#7
In reply to #2

Re: Load Cell Drift in Scales

04/27/2010 3:44 AM

Not quite:

I make it about 0.0003 g for a change of 20 millibars (100 gram weight).

__________________
If you spend all your time looking for people and things to complain about: trust me, you will find plenty to complain about.
Register to Reply
Active Contributor

Join Date: Aug 2007
Posts: 21
#3

Re: Load Cell Drift in Scales

04/26/2010 10:24 AM

100g capacity, 0.0001 resolution, 0.0006g tolerance.

Register to Reply
Anonymous Poster
#4
In reply to #3

Re: Load Cell Drift in Scales

04/26/2010 10:49 AM

As long as the scale is enclosed, (and I am assuming that it is), it is not normal for it to be that far off.

Register to Reply
Guru
Hobbies - Fishing - Old Salt Hobbies - CNC - New Member United States - US - Statue of Liberty - New Member

Join Date: Mar 2007
Location: Rosedale, Maryland USA
Posts: 5197
Good Answers: 266
#5

Re: Load Cell Drift in Scales

04/26/2010 1:49 PM

How old is the scale?

__________________
Life is not a journey to the grave with the intention of arriving in a pretty, pristine body but rather to come sliding in sideways, all used up and exclaiming, "Wow, what a ride!"
Register to Reply
Guru

Join Date: Dec 2006
Location: Germany 49° 26' N, 7° 46' O
Posts: 1950
Good Answers: 109
#6

Re: Load Cell Drift in Scales

04/27/2010 2:11 AM

Hi,

if you are sure about the temperature to be ok and the voltage too, then your scale very likely was overloaded or shocked.

Mechanical overload will cause permanent strain in the flexures that suspend the weight.

This permanent strain is very temperature dependent (in principle near -0.03%/degreeC) but in reality amplified by the principles of the flexures.

You can check this if you make temperature tests with a. no load, b. full load in the center and c. full load in one of the 4 corners. So with these 6 tests in a full temperature cycle you will get either all data within specs. If so your mechanical system of the scale was not overloaded. If not then very likely you had an overload.

Maybe you can relief the stress-strain damage a bit by storing the scale at the highest allowed temperature for some weeks.

Another possibility is that your scale was overloaded at transport. Anti-shock packaging is not at all simple.

If the temp-tests fail then you need a new scale.

Have success

RHABE

Register to Reply
Active Contributor

Join Date: Aug 2007
Posts: 21
#8
In reply to #6

Re: Load Cell Drift in Scales

04/27/2010 8:57 AM

Thanks.

More facts: 1. It is three years old. 2.The first and second calibration was good. 3.It was calibrated with the shield. 4. It failed at 70,80 90, 100g.The tech did not pay close attention to the temp, but I monitored the temp and it was 69F/ 36 RH. The report has a check mark next to " ACCEPTIBLE".

This scale had not been used for a year, but could have been overloaded one of the last times it was used, but I cannot verify this.

thanks again

Register to Reply
Guru

Join Date: Oct 2006
Location: Surrey BC Canada
Posts: 1571
Good Answers: 42
#9
In reply to #8

Re: Load Cell Drift in Scales

04/27/2010 10:19 PM

Retry with fresh batteries and known temperature.

After a year out of service I would suspect the power supply and temperature.

(For the good old days of sensitive balance beam scales! 35 years ago we had some marvelous ones at university all enclosed in a glass and wood box so air drafts didn't tip the balance! The scale was used in drying miniscule chemical samples to constant weight.)

Register to Reply
Active Contributor

Join Date: Aug 2007
Posts: 21
#10
In reply to #9

Re: Load Cell Drift in Scales

04/28/2010 7:31 AM

Thanks,

The calibration was done in a enclosed room with the shield installed, and an AC adaptor.

Register to Reply
Register to Reply 10 comments

"Almost" Good Answers:

Check out these comments that don't yet have enough votes to be "official" good answers and, if you agree with them, vote them!
Copy to Clipboard

Users who posted comments:

Anonymous Poster (2); GW (1); need_to-Know (3); ozzb (1); PWSlack (1); Randall (1); RHABE (1)