how to calibrate temperature controller?
Re: Temperature Controller Calibration
02/18/2011 4:16 AM
If all else fails, read the instructions.
02/18/2011 4:28 AM
Duplicate thread alert!
02/18/2011 4:33 AM
Inadequate thread alert!
Good Answers: 11
02/18/2011 5:12 AM
http://www.omega.com/prodinfo/calibrators.html
Read and be logically imaginative.
02/18/2011 12:42 PM
Hot plate and beaker of water for high temp.
Ice bath for low.
02/23/2011 7:05 AM
1. First check which type of sensor you are using with your temp. controller. i.e. RTD or Themocouple
2. Then connect calibrator probes to the sensor terminals of your temp. controller.
3. power on both calibrator & your temp. controller.
4. Select the sensor type on the calibrator. i.e. RTD or Thermocouple.
5. Feed the values to the temp. controller through calibrator from 0 to max. range of the sensor selected & vice versa.
6. Cross check both values on calibrator & temp. controller whether the values shown on temp. controller are within acceptance limit.