Dear All,
The plant that i am working we have a quality control dept, we are facing a problem of frequent equipment failure. To brief, this dept is supplied from two 30KVA UPS of Emerson Liebert. both are 3 Phase, 4 wire system and paralleled. and the UPS load consists of equipments like HPLC, GC, weighing balances and other analytical instruments only
Very recently we observed some black streaks on a weighing balance and a pH meter in this lab. Both equipment vendors suspected there is an issue with power fluctuations and grounding issue. When checked by UPS vendor he said there is no issue with his UPS and outputs and earthing are fine
I have checked and found that the UPS available is ample, and there is no chance for any power interruption. I have checked the supply voltage and earth and found supply voltages to be with in limits and earth to neutral voltage is less than 3 V.
We had a similar kind of failure in weighing balance about 3 months back where in the equipment manufacturer suggested that the issue is with UPS power.
A recent study of our power system showed that the harmonic distortion(THD) is less than 5%. I am unable to find what else could be the reason for this frequent failure of equipments
Thanks and regards