Mass Flow Meter Installation

06/14/2013 8:21 AM

Dear All
We are having one application where we are using Mass Flowmeter of Siemens Make MAS 2100 with 1/4" NPT process connection.
Line pressure is 20 Kg/cm2.
We shall be using 1" Flange(300#) with 1/4" NPT(M) connection.
My question is is it reliable to use thread connection with such a big pressure in the line?
What other options are available for Mass flow Meter installation?

Register to Reply
Pathfinder Tags: mass flowmeter
Interested in this topic? By joining CR4 you can "subscribe" to
this discussion and receive notification when new comments are added.
Guru

Join Date: Oct 2008
Posts: 42355
Good Answers: 1693
#1

Re: Mass Flow Meter Installation

06/14/2013 8:54 AM

What material is the pipe? Steel, aluminum, plastic, titanium, lead, copper?

What does the piping specification say?

What does your piping engineer say?

Process engineer?

Not to mention the previous questions?

Register to Reply
Guru

Join Date: Jul 2010
Posts: 669
Good Answers: 176
#2

Re: Mass Flow Meter Installation

06/14/2013 9:11 AM

The spec sheet says

Register to Reply
Guru
Engineering Fields - Mechanical Engineering - New Member India - Member - New Member

Join Date: Aug 2008
Location: City of destiny, INDIA
Posts: 775
Good Answers: 67
#3

Re: Mass Flow Meter Installation

06/15/2013 6:06 AM

What fluid, gas or liquid?

Any way 20 kg/cm2 is not a so high pressure that threaded connection can not be used. Only thing to be ensured that all the fittings and sealants used shall be designed for the pressure, temp and fluid intended.

Register to Reply
Guru
Engineering Fields - Instrumentation Engineering - EE from the the Wilds of Pa.

Join Date: Feb 2006
Location: middle of nowhere, Pennsylvania
Posts: 2603
Good Answers: 63
#4

Re: Mass Flow Meter Installation

06/17/2013 8:50 AM

Your pressure converts to 284 psi and you have 300 psi fittings - that's okay. Now as far as threaded fittings go, when I was a well analyst many years ago we had high pressure "hammer" unions for most fittings, but some still had to be NPT threads. On our control standpipe on 4000 psi gas wells, once properly tightened, we had no leakage problems, and if something is going to leak, it is high pressure natural gas. The gas companies just can't seem to get a nice slow pressure up on a standpipe - a quick hit of high pressure finds leaks really fast. Yes, we had to tighten some joints occasionally, but the threads held when snugged in tight.

__________________