HF Leakage of Electrosurgical Unit

09/25/2013 5:12 AM

Basically I am an electrical engineer and I don't have much experience in bio-medical field. Can somebody give an idea about what is doing in HF leakage test of electrosurgical unit and what parameter is measuring in that test ?

I heard that it is something related to the property of current to stray at high frequencies. But I don't understand what is meant by that.

Can somebody help me please...

__________________
If you can't find a way, make one........
Register to Reply
Interested in this topic? By joining CR4 you can "subscribe" to
this discussion and receive notification when new comments are added.

"Almost" Good Answers:

Check out these comments that don't yet have enough votes to be "official" good answers and, if you agree with them, vote them!
Guru
United Kingdom - Member - Indeterminate Engineering Fields - Control Engineering - New Member

Join Date: Jan 2007
Location: In the bothy, 7 chains down the line from Dodman's Lane level crossing, in the nation formerly known as Great Britain. Kettle's on.
Posts: 32175
Good Answers: 839
#1

Re: HF leakage of electrosurgical unit

09/25/2013 5:19 AM
__________________
"Did you get my e-mail?" - "The biggest problem in communication is the illusion that it has taken place" - George Bernard Shaw, 1856
Register to Reply
Power-User
Engineering Fields - Electrical Engineering - New Member Engineering Fields - Power Engineering - New Member

Join Date: Mar 2012
Location: the sandbox
Posts: 341
Good Answers: 6
#2

Re: HF Leakage of Electrosurgical Unit

09/26/2013 6:46 AM

Please provide a reference to this requirement. (link)

__________________
BSEE but always learning
Register to Reply
Guru

Join Date: May 2010
Location: Metro.Manila, Philippines.
Posts: 1271
Good Answers: 27
#3

Re: HF Leakage of Electrosurgical Unit

09/26/2013 9:25 AM

Basically an ESU is a very high frequency current generator, where the generated frequency is high enough beyond the threshold that will cause the muscle twitching sensation on the patient. It is use for cutting tissues, coagulating blood and/or a combination thereof. Being a high frequency generator, it behaves like any other HF or VHF transmitter that if an AM radio receiver is close by, its reception will be over-ridden by the ESU generated current!

I believe what you may be looking for is how well and effective the designed current return path of that ESU unit? Normally the output of the generator beside being measured, calibrated and monitored, it also imperative that whatever that power was outputted, it should return back to the generator to complete that circuit.

An ESU unit if properly designed and functioning, monitors and measures the current that leaves and comes or collected back to the unit. This approach not only ensures system effectiveness, and by monitoring the return path it also ensures that any stray currents are minimized thus enhancing electrical safety. Any imbalance monitored triggers corresponding alarm parameters!

__________________
vsar
Register to Reply
Power-User
India - Member - Justin Anto Engineering Fields - Electrical Engineering - Marks never made me laugh, but good memories did

Join Date: Sep 2010
Location: Thrissur, Kerala, India
Posts: 177
Good Answers: 3
#5
In reply to #3

Re: HF Leakage of Electrosurgical Unit

10/11/2013 4:56 AM

That means, by doing HF leakage test, we are monitoring the leakage of power, right ?

Or are are monitoring the leakage current ?

__________________
If you can't find a way, make one........
Register to Reply
Guru

Join Date: May 2010
Location: Metro.Manila, Philippines.
Posts: 1271
Good Answers: 27
#6
In reply to #5

Re: HF Leakage of Electrosurgical Unit

10/11/2013 8:22 AM

When you do that HF leakage test, basically you are trying to verify that whatever monitored power that comes out (delivered to the surgical site), effectively returned and find its direct path back to the machine to complete its circuit!

Bear in mind that power is a combination of current and voltage! One is a function of the other and therefore cannot be separated!

__________________
vsar
Register to Reply Score 1 for Good Answer
Guru

Join Date: May 2010
Location: Metro.Manila, Philippines.
Posts: 1271
Good Answers: 27
#4

Re: HF Leakage of Electrosurgical Unit

09/26/2013 9:39 AM

Cont'd

On some ESU models, an alarm feature called REM, (return electrode monitoring), are employed. This measures the continuity and skin impedance of the return electrode circuitry. It helps guide the proper placement of the grounding or return electrode close to the surgical site, insuring the shortest current return path that also enhances the expected cutting results using the least or minimum power setting!

Good luck

__________________
vsar
Register to Reply
Power-User
Engineering Fields - Electrical Engineering - New Member Engineering Fields - Power Engineering - New Member

Join Date: Mar 2012
Location: the sandbox
Posts: 341
Good Answers: 6
#7

Re: HF Leakage of Electrosurgical Unit

10/11/2013 11:09 AM

Check the link given above for Ultramedic. That device monitors 30hz - 2.5Mhz under load, but since outputs below 100hz are filtered out, those outputs should be zero.

So basically the intent of the device is:

1. monitor the intended RF output

2. monitor the 50 or 60 hz current which should be filtered out

3. monitor all the harmonics of that 50 or 60hz current.

NEC requires an alarm to sound for total hazard currents above 5mA.

__________________
BSEE but always learning
Register to Reply
Guru

Join Date: May 2010
Location: Metro.Manila, Philippines.
Posts: 1271
Good Answers: 27
#8
In reply to #7

Re: HF Leakage of Electrosurgical Unit

10/11/2013 12:32 PM

For Patient Safety;

There are two types of current leakages that are present, being minimized and continuously monitored in this particular type of medical device!

1) Regular low frequency or power line leakage current (either 50 or 60 Hz) derived from or based on power source. This determines the micro-shock hazard threshold of the unit.

2) High frequency current leakages developed from the generator cutting/coagulating currents. This determine the amount of stray current that can cause un-intended damage to the surrounding tissues or organs in the patient's body!

__________________
vsar
Register to Reply
Register to Reply 8 comments

"Almost" Good Answers:

Check out these comments that don't yet have enough votes to be "official" good answers and, if you agree with them, vote them!
Copy to Clipboard

Users who posted comments:

1capybara (2); justinanto1987 (1); PWSlack (1); vsar (4)

Previous in Forum: Goldfish Tumor?   Next in Forum: Flesh Eating Drug Hits US

Advertisement