High Vacuum Systems

06/20/2010 10:00 AM

I started to work in high vacuum systems. I wondered how long it would take to reach the required vacuum of 1e-6 torr. I was told that for low vacuum it can be calculated but for high vacuum it depends mainly on the surface area and the outgassing of each surface (not very practical in complex systems). Is there any simple way to get a rough estimate on how long it will take?

Register to Reply
Interested in this topic? By joining CR4 you can "subscribe" to
this discussion and receive notification when new comments are added.

Good Answers:

These comments received enough positive votes to make them "good answers".
Guru

Join Date: Oct 2008
Posts: 42355
Good Answers: 1693
#1

Re: high vacuum systems

06/20/2010 10:16 AM

Vacuum - Evacuation Time

Listen to the old timers.

Register to Reply
4
Guru
Engineering Fields - Electrical Engineering - Been there, done that. Engineering Fields - Control Engineering - New Member

Join Date: Dec 2008
Location: Long Island NY
Posts: 15602
Good Answers: 982
#2

Re: high Vacuum Systems

06/20/2010 1:51 PM

Yes, high vacuum systems will be dominated by out gassing concerns that prevent time predictions. After the standard low pressure equations give you a rough idea how quickly it will take to get most of the already gaseous materials out of a chamber the physics and chemistry will change in may ways. A good paper titled "High Vacuum Gas Pumping and Boundary Coupling" show many of the complications that occur in pumping a system down to these low vacuum conditions. In this paper you will notice that the predicted particle trajectory path towards the pump port (turbo or ion) becomes more important than any relative pressure differentials. Now out gassing complicates this even further by continually adding new molecules to the chamber as those that enter the pumping port get removed. One greasy sweaty finger print on a Ultra High Vacuum (1e-9) chamber surface can add hours to a pump down. A few unseen insects can add days.

An added out gassing complication that surprises many people is the trapped volume problem that most hardware can produce. Yes, all hardware used inside a high vacuum chamber must be throughly cleaned of all machining lubricants used in fabrication. Now think about what happens with a blind tapped hole filled with a clean bolt tightened into position. At the tip of the bolt there is now a reservoir of gas that will slowly leak past the gap found between the threads into the chamber. To speed up this trapped gas extraction, often the bolts used in what must be a tapped blind hole will have a tiny hole drilled down the shaft center or a groove cut along the length of the threads to vent the gas. Compounding this complication is that any two surface areas that meet each other unless they are both perfectly flat can trap a tiny volume of gas that will slowly leak out. So while the larger reservoirs of blind holes can be both quantified and mitigated, the remaining hardware related trapped gas cannot be quantified or mitigated.

__________________
"Don't disturb my circles." translation of Archimedes last words
Register to Reply Good Answer (Score 4)
Guru
Hobbies - Musician - Engineering Fields - Chemical Engineering - New Member Engineering Fields - Control Engineering - New Member Engineering Fields - Instrumentation Engineering - New Member

Join Date: Jan 2007
Location: Moses Lake, WA, USA, Thulcandra - The Silent Planet (C.S. Lewis)
Posts: 4216
Good Answers: 194
#4
In reply to #2

Re: high Vacuum Systems

06/20/2010 10:30 PM

Dang! VGA! Your experience in this area is quite evident. That, along with your detailed reply, is the essence of what we do here. Well done.

__________________
"Reason is not automatic. Those who deny it cannot be conquered by it. Do not count on them. Leave them alone." - Ayn Rand
Register to Reply
Guru
United Kingdom - Member - Old New Member

Join Date: Jul 2005
Location: South east U.K.
Posts: 3695
Good Answers: 93
#6
In reply to #2

Re: high Vacuum Systems

06/21/2010 5:45 AM

Good comments from Redfred. We spend a lot of time trying to avoid trapped volumes in vacuum assemblies.

I've been told that leaving a finger print on a component in a vacuum chamber will increase the pumping time by 1 hour. I have no way of knowing how true this is. We pump down to 1x10-10 torr so our pumping cycle is quite long.

__________________
I didn't have a really important life, but at least it's been funny (Lemmy Kilminster 1945-2015)
Register to Reply
Guru

Join Date: Oct 2008
Posts: 42355
Good Answers: 1693
#3

Re: High Vacuum Systems

06/20/2010 2:07 PM

I don't know what you are doing, but, this:

Outgassing Data for Selecting Spacecraft Materials System

was my bible for 20 years when working for a major builder of military/spy satellites.

Cheers.

Register to Reply
Guru

Join Date: Dec 2006
Location: Germany 49° 26' N, 7° 46' O
Posts: 1950
Good Answers: 109
#5

Re: High Vacuum Systems

06/21/2010 3:21 AM

Hi,

some basics:

Glass is best, but often not allowed. Surface can be very clean and is usually very smooth.

Stainless steel, other non-corrosion alloys as based on Nickel, Titanium are very good if the surface is smooth and if there are no included voids (as mentioned above below screws there are these trapped gas leaks from machining of surfaces and defects in materials.

Avoid any brass: the Zinc will do no good, allow copper or copper-alloys only where necessary the oxide will absorb big amounts of gas.

If working with aluminum then no 7xxx-alloys (Zn) but 2xxx or 6xxx are fine. Do the final milling without any lubricant or with isopropyl (fire and explosion hazard). Do not allow any waxy or fatty stuff ever touch the aluminum as these may migrate in and through aluminum. Add a very thin (much below 1 µm) very special anodising layer to preserve initial quality. Or make a thermal treatment in 5%-Oxygen-Argon mixture.

Use only the shortest screws that will do the fastening and drill lengthwise as thick a hole as possible.

Any non-porous ceramic is a very good choice but any porous ceramics - magnets for example are often porous - may be constant sources of gas for years. I once measured ball-bearing-oils, magnet-rings, insulating materials for copper-wirings and more and the magnets stayed at high outgassing rates for months and never came down to acceptable levels.

Any plastics to be avoided as far as possible, choice according to the above-mentioned NASA or newer ESA specifications, your own out outgassing procedure a necessity (24 hours at highest allowed temperature or minimum 30 °C above highest service temperature). Never use any poly-amides (Nylon).

Any glues should be hot-curing and carefully mixed to ensure total reaction - ask the chemical specialists of the glue manufacturer if the 2 components of the glue will totally react with each other or if the excess by non-exact dosing will remain non-cured and slowly comes out outgassing into your system, may be cracked by any ions and then be able to quickly migrate to everywhere.

Inspect and rework your systems flanges to achieve and maintain a good polishing finish if using VITON-O-rings. Use double Viton-seals with in-between pumping if one is not good enough or use cooled Viton-seals.

I don't like the copper (CONFLAT) seal rings as these are expensive and are requiring expensive flanges. I like better self-made seal-rings from silver (Ag 999 to AgCu20 is fine), with some training you can easily braze these with a gas-torch. To be put either between flat flanges or one side flat the other side with a V-groove. To be compressed to roughly 70% of original diameter of wire depending on quality of the seal surfaces.

All this done and never working without cotton-gloves you will pump down a chamber of 10 liter volume and 0.5m² surface and 2m total length of Viton-seals by a 200l/s turbo-pump backed by a 16m³/h roughing pump within 1 hour to 10-6 mbar.

The Viton-seals were the biggest leaks in this chamber, these rings leak and store gas!

If you change the Viton to Silver then you will reduce pumping time to below 10 minutes. The tiniest scratch on a flange will make everything much worse!

Be patient with your system and have success

RHABE

Register to Reply
Guru
Hobbies - DIY Welding - New Member

Join Date: Sep 2007
Posts: 1601
Good Answers: 58
#7

Re: High Vacuum Systems

06/21/2010 8:13 AM

Although others have provided significant detail, it goes without saying that you must eliminate all leaks. Not only must direct leak paths be eliminated, indirect paths, called virtual leaks must be dealt with. An example of a virtual leak is a drilled and tapped bottoming hole with a screw in it. The gas volume under the screw may not leak to atmosphere, but represents a gas source that will leak into the pumped chamber for a long time. The solution is to drill a vent hole so this trapped gas can be pumped directly.

Register to Reply
Power-User

Join Date: Mar 2008
Location: MA 01864, USA
Posts: 453
Good Answers: 7
#8

Re: High Vacuum Systems

06/21/2010 8:39 AM

This is tough question. I work routinely at vacuum stability of better than E-9. It is the surface desorption behavior, along with grease, dead areas and helium which we use to test the vacuum stability absorption properties with the wall of stainless steel used to generate vacuum. One leak part and will take 24 hours before we start test again. this is to complicated topic to put in one para .

Careful with measurement and stability of your vacuum tester along with the material system and design

__________________
Masyood
Register to Reply
Register to Reply 8 comments