We recently purchased a centrifugal transfer pump to supply water to a laboratory upstairs on second floor of main building from a condensate storage tank in our plant, outside the building. The supply off this tank is a 6" pipe, so no problems at all transfering lab water. The tank rests approximately 1 ft off zero level of floor, and the pump rests similarly, inside the building in the ground floor, ~ 1 ft off the floor. Tank is 24 feet tall side wall, and the lab is approximately 20ft above ground floor to the floor of the lab.
I am working on setting up this magnetically coupled pump "Little Giant" #4-SD-SC with a pressure switch to control run/stop of the motor, based upon if the valve in the laboratory is open or is closed. Thus far I have 14 psi set as cut-out pressure, and 10 psi set as cut-in pressure. I suspect these numbers are perhaps a tiny bit too high, as the pump may not have any output past 24 ft differential head. Under higher settings I tried first, the pump didn't want to cut-out at all
. 14 psi = 32.3 ft H2O, and 10 psi = 23.1 ft H2O. By the way, the water will flow freely without the pump being energized when tank is full. Water flow under this condition is approximately 0.2 L/m, and with pump energized this increases to 0.5 l/m on a stainless steel 3/8" tubing line approximately 100 ft length.
Any general ideas to improve this set-up (I wanted to install a manual switch, but I&E has been particularly "lazy" of late with respect to my various projects)?