Liquid Level Measurement in Agitated Tanks

07/17/2013 1:32 AM

Guys,

We have to offer a level sensing mechanism for corrossive liquids in agaitated tanks. The customer will use load cells as one set of measurement and then use another type to cross check and validate the acual volumes. Some type of API where the vloumes are super critical - so we are told.

We looked at various methods and devices available including radar. The radar type sensor, we understand from some of the vendors will give an accuracy of+/- 3 mm. The customer is looking for tighter accuracies - like +/- 1 mm.

What suggestions do you have? Volumes of tanks (there are several) range from 250 l to 1000 l.

Thanks a ton! (or a mm!)

__________________
Ashok Pillai
Register to Reply
Interested in this topic? By joining CR4 you can "subscribe" to
this discussion and receive notification when new comments are added.

Good Answers:

These comments received enough positive votes to make them "good answers".

"Almost" Good Answers:

Check out these comments that don't yet have enough votes to be "official" good answers and, if you agree with them, vote them!
Guru
Technical Fields - Technical Writing - New Member Engineering Fields - Piping Design Engineering - New Member

Join Date: May 2009
Location: Richland, WA, USA
Posts: 21017
Good Answers: 795
#1

Re: Liquid level measurement in agitated tanks

07/17/2013 3:05 AM

Your customer wants ±1mm accuracy in an agitated tank? That's completely crazy.

__________________
In vino veritas; in cervisia carmen; in aqua E. coli.
Register to Reply
Associate

Join Date: Apr 2007
Posts: 45
#2
In reply to #1

Re: Liquid level measurement in agitated tanks

07/17/2013 3:15 AM

Tell me about it! Well, the radar guys have software that takes into account the readings from the agitated mass and gives a reading within +/- 3 mm. But that's all they are willing to guarantee in agitated as well as in still conditions.

We could give a SOP to the customer that says - stop the agitation and uses this gizmo - you get the level with accuracy of +/- 1 mm. Question - what gizmo?

Alternately I was thinking if we can measure very accurately the flows into the vessel (coriolis flow meter? don't know any thing about it - just looked it up on the net).

__________________
Ashok Pillai
Register to Reply
Guru
United Kingdom - Member - Indeterminate Engineering Fields - Control Engineering - New Member

Join Date: Jan 2007
Location: In the bothy, 7 chains down the line from Dodman's Lane level crossing, in the nation formerly known as Great Britain. Kettle's on.
Posts: 32175
Good Answers: 839
#4
In reply to #2

Re: Liquid level measurement in agitated tanks

07/17/2013 3:50 AM

Load cells and a density transmitter seems to be the way to go. Not knowing the liquid, the temperature or the pressure prevents further suggestions being made.

Line diameters and flowrates would be valuable too.

Are these vented tanks?

Is there an inert gas blanket?

__________________
"Did you get my e-mail?" - "The biggest problem in communication is the illusion that it has taken place" - George Bernard Shaw, 1856
Register to Reply
Guru
United Kingdom - Member - Indeterminate Engineering Fields - Control Engineering - New Member

Join Date: Jan 2007
Location: In the bothy, 7 chains down the line from Dodman's Lane level crossing, in the nation formerly known as Great Britain. Kettle's on.
Posts: 32175
Good Answers: 839
#3

Re: Liquid level measurement in agitated tanks

07/17/2013 3:49 AM

+/- 1mm? At what temperature swing and what density swing?

Define <...corrossive liquids...>. All liquids are corrosive to some materials; one wouldn't put hot water in a vessel made of sucrose, for example.

__________________
"Did you get my e-mail?" - "The biggest problem in communication is the illusion that it has taken place" - George Bernard Shaw, 1856
Register to Reply
Associate

Join Date: Apr 2007
Posts: 45
#5
In reply to #3

Re: Liquid level measurement in agitated tanks

07/17/2013 4:57 AM

Guru,

Right! alwayes hated using hot water in those fancy sucrose vessels!

These are vessels in SS or glasslined

The operating parameters :

Temp - Room temperture ( 30 deg C) ro max 80 degree C
Operating pressure - 3 bar
Process constituents - solvents + WFI + product mass

Ph - mildly Alkaline: in one case acidic where it is a glasslined vessel

Do not have info on density mass but let me know how it could be relevant and I will ask.

There is no nitrogen blanketing, tanks are not vented to atmosphere.

__________________
Ashok Pillai
Register to Reply
4
Guru
United Kingdom - Member - Indeterminate Engineering Fields - Control Engineering - New Member

Join Date: Jan 2007
Location: In the bothy, 7 chains down the line from Dodman's Lane level crossing, in the nation formerly known as Great Britain. Kettle's on.
Posts: 32175
Good Answers: 839
#6
In reply to #5

Re: Liquid level measurement in agitated tanks

07/17/2013 5:41 AM

Acronymectomy for other readers: WFI means "water for injection" - pyrogen-free water.

Here is the deal:

  • Don't use capacitance if the composition of the product changes.
  • Don't use ultrasonic if there is froth or an agitated surface, or a risk of droplets condensing on the sensor or if the pressure is not ambient.
  • Don't use radar if there is an agitated surface.
  • Displacement and back-pressure won't give +/- 1mm, nor will RF admittance.
  • Submerged differential pressure won't work if there is agitation or the density changes.

It's starting to look like load cells under each vessel to give the mass of the material and a density meter on the outlet. Divide mass by density and one gets volume. Doing it this way means the temperature is irrelevant as density is a function of temperature; if the temperature swings in the vessel, then so will the volume, which may be significant.

Good luck.

<unsubscribes>

__________________
"Did you get my e-mail?" - "The biggest problem in communication is the illusion that it has taken place" - George Bernard Shaw, 1856
Register to Reply Good Answer (Score 4)
Associate

Join Date: Apr 2007
Posts: 45
#7
In reply to #6

Re: Liquid level measurement in agitated tanks

07/17/2013 5:54 AM

Terrific! Thanks a lot. ...

__________________
Ashok Pillai
Register to Reply
Guru

Join Date: Oct 2008
Posts: 42355
Good Answers: 1693
#10
In reply to #7

Re: Liquid level measurement in agitated tanks

07/17/2013 7:07 PM

You have not defined "agitation".

Register to Reply
Associate

Join Date: Apr 2007
Posts: 45
#19
In reply to #6

Re: Liquid level measurement in agitated tanks

07/18/2013 7:34 AM

HI PWSlack,

A bit more on your recomendations: "Don't use ultrasonic if there is froth or an agitated surface, or a risk of droplets condensing on the sensor or if the pressure is not ambient."

Question is why? Maybe you have experience in using ultrasonic sensors but I wanted to check with you a device I located on the net - www.solidat.com. One of the products featured, GaugerGSM claims .."Embedded algorithms extract false echoes, compensate for temperature and handle turbulent targets".

Would also appreciate letting me know why pressure different from ambient could give false readings? I can guess droplets interfering with the reading of the reflected waves...

Thanks again

__________________
Ashok Pillai
Register to Reply
Guru
United Kingdom - Member - Indeterminate Engineering Fields - Control Engineering - New Member

Join Date: Jan 2007
Location: In the bothy, 7 chains down the line from Dodman's Lane level crossing, in the nation formerly known as Great Britain. Kettle's on.
Posts: 32175
Good Answers: 839
#30
In reply to #19

Re: Liquid level measurement in agitated tanks

07/24/2013 5:02 AM

<Resubscribes>

<...recomendations: "Don't use ultrasonic if there is froth or an agitated surface, or a risk of droplets condensing on the sensor or if the pressure is not ambient."...>

'Cos it won't work!

<unsubscribes>

__________________
"Did you get my e-mail?" - "The biggest problem in communication is the illusion that it has taken place" - George Bernard Shaw, 1856
Register to Reply
3
Guru
Engineering Fields - Instrumentation Engineering - EE from the the Wilds of Pa.

Join Date: Feb 2006
Location: middle of nowhere, Pennsylvania
Posts: 2603
Good Answers: 63
#8

Re: Liquid level measurement in agitated tanks

07/17/2013 8:13 AM

Well you need to learn, as I did, that every now and then the customer is wrong. Many years ago, when I was a new controls engineer, I spent a month trying to get results for a customer who had requested impossible results. In the end he accepted what was possible in the real world, but in the mean time, I was having very sleepless nights, and my company management was hounding me daily to get the project out the door.

For a surface that is constantly rippling, the specs are way too tight.

__________________
Remember when reading my post: (-1)^½ m (2)^½
Register to Reply Good Answer (Score 3)
2
Guru

Join Date: Jul 2010
Posts: 669
Good Answers: 176
#9

Re: Liquid Level Measurement in Agitated Tanks

07/17/2013 10:50 AM

The reason for duplicity of measurement is the problem load cells have with agitation and their only solution is to average out the readings.

There's one exception - Sartorius. Do yourself a favor and dheck out the one minute 'bonus' video (scroll down half way) here

http://www.lesman.com/train/webinars/Webinar_SartoriusTankHopperWeighing.htm

to see how their vessel mounting devices eliminate the forces contributed by agitation. Their matched load cells and mounting really does a great job.

Connect a DP transmitter with a flex hose at the bottom of the tank and average the heck out of its readings as a secondary check. But the Sartorius will be dead nuts on, regardless.

Register to Reply Good Answer (Score 2)
Associate

Join Date: Apr 2007
Posts: 45
#27
In reply to #9

Re: Liquid Level Measurement in Agitated Tanks

07/22/2013 8:08 AM

Thanks Iris. This is good. Sartorius is the gold standard for load cells and if we can use a density meter as suggested by PW Slack it may be good enough....

__________________
Ashok Pillai
Register to Reply
Guru
Engineering Fields - Mechanical Engineering - New Member

Join Date: May 2008
Location: CHENNAI, TAMIL NADU, INDIA.
Posts: 1851
Good Answers: 65
#11

Re: Liquid Level Measurement in Agitated Tanks

07/17/2013 11:52 PM

Dear Mr.Ash55,

As I understand from your statement, agitating vessel means LEVEL CANNOT BE MEASURED CORRECTLY.

If so, you can provide a gauge glass, outside the vessel, like as in Boiler, where the liquid will reach its level and stay undisturbed, and level sensor can be clubbed to this. Perhaps some working difficulty may come and to be overcome.

This applies to cylindrical vessel only -vertically mounted. Let us see responses from CR4 Members.

DHAYANANDHAN.S

Register to Reply
Guru

Join Date: Dec 2010
Location: North West England
Posts: 1170
Good Answers: 153
#12

Re: Liquid Level Measurement in Agitated Tanks

07/18/2013 3:48 AM

If you cannot measure it while it is being agitated, measure it when it is not being agitated. Meter the liquid as it put into, and taken out of the tank, and the difference is what is in the tank. You will need positive displacement flowmeters to get that sort of accuracy.

Register to Reply Score 1 for Good Answer
Guru
Engineering Fields - Mechanical Engineering - New Member

Join Date: May 2008
Location: CHENNAI, TAMIL NADU, INDIA.
Posts: 1851
Good Answers: 65
#14
In reply to #12

Re: Liquid Level Measurement in Agitated Tanks

07/18/2013 5:17 AM

Dear Mr.jhhassociates

Your view may not work for contiuous flow and agitation. It will be OK for batch type filling, drain/discharge after agitation.

Pl. let us know from you on the continuos flow/agitation.

DHAYANANADHAN.S

Register to Reply
Associate

Join Date: Mar 2013
Location: North East UK
Posts: 27
Good Answers: 2
#13

Re: Liquid Level Measurement in Agitated Tanks

07/18/2013 4:07 AM

Hi Ash,

I work in a facility that uses agitated tanks between 250L and 10,000L we use load cells and a calibrated sight tube on the outside of the tank. This works for everything from WFI to Buffers containing ethanol / NaOH etc.

Aslong as the inprocess testing is adequate to prove the contents are the correct concentration / ratio etc then there should be no problems.

Register to Reply
Associate

Join Date: Apr 2007
Posts: 45
#15
In reply to #13

Re: Liquid Level Measurement in Agitated Tanks

07/18/2013 5:33 AM

Hi only1yeti,

Yes, there are graduated view glasses on the side. Two problems with it:

1. It cant show the level all the way down and up (sight glasses on side of vessel as built now).

2. Even if such extreme levels are not required to be met, the customer needs the data to be electronically be available from two sources and verified electronically (CFR 21). So we will need the level markings to be picked up and converted to a level electronically.....

__________________
Ashok Pillai
Register to Reply
2
Associate

Join Date: Mar 2013
Location: North East UK
Posts: 27
Good Answers: 2
#16
In reply to #15

Re: Liquid Level Measurement in Agitated Tanks

07/18/2013 5:37 AM

aslong as the paperwork is correct and the sight tube measurements are verified by a 2nd operative then it meets GMP requirements for production of medical devices/ Drug GMP (Annex 1 eudralax), CFR 21 is only required on electronic devices so aslong as its non electronic it doesnt have to conform. as for the sight tubes not going all the way down that is where different sized vessels and validation normally comes into play.

Register to Reply Good Answer (Score 2)
Associate

Join Date: Apr 2007
Posts: 45
#17
In reply to #16

Re: Liquid Level Measurement in Agitated Tanks

07/18/2013 5:41 AM

Thanks, that's a very interesting piece of information that we can use when discussing with the customer... do you know where I could get hold of the relevant extract that allows it?

__________________
Ashok Pillai
Register to Reply
Associate

Join Date: Mar 2013
Location: North East UK
Posts: 27
Good Answers: 2
#18
In reply to #17

Re: Liquid Level Measurement in Agitated Tanks

07/18/2013 5:50 AM

try googling the Eudralex (european Drug / Medical Device GMP requirements) that is prob a good place to start in a highly regulated industry like this.

Register to Reply Score 1 for Good Answer
Associate

Join Date: Mar 2013
Location: North East UK
Posts: 27
Good Answers: 2
#23
In reply to #17

Re: Liquid Level Measurement in Agitated Tanks

07/18/2013 11:51 AM

for the rules of CFR 21 part 11 go here --> http://www.accessdata.fda.gov/scripts/cdrh/cfdocs/cfcfr/CFRSearch.cfm?CFRPart=11

for the Eudralex material --> http://ec.europa.eu/health/documents/eudralex/ is your best bet.

Register to Reply
Associate

Join Date: Apr 2007
Posts: 45
#26
In reply to #13

Re: Liquid Level Measurement in Agitated Tanks

07/22/2013 8:06 AM

You know if there was some way the calibrated sight glass on the side of the vessel could be read and validated with the load cell readings electronically, we may have a solution.

I was actually thinking of having webcams that will take a shot of the level through the calibrated sight glass as a way of validating the load cell reading. So, although the validation is not auto, there is an electronic record of the time when the load cell reading was done and the cam shot.

Got to talk to the customer in a couple of days with some solutions.

What do you think -dumb idea?

__________________
Ashok Pillai
Register to Reply
Power-User

Join Date: Jan 2010
Location: India
Posts: 166
Good Answers: 1
#20

Re: Liquid Level Measurement in Agitated Tanks

07/18/2013 8:02 AM

Have a glass level gauge beside the tank. Measure level here. The level averages out nicely. You are likely to get much better reading and tolerance.

Register to Reply
Guru

Join Date: Nov 2007
Location: Eastern Arizona mountains on Route 666 about a mile from God's country
Posts: 1676
Good Answers: 122
#21

Re: Liquid Level Measurement in Agitated Tanks

07/18/2013 10:25 AM

Here is a link to ER guided wave tank level monitoring that is at .5mm accuracy.

We use the 5300 unit in our agitated tanks and have had excellent results with no failures since installation 5 years ago.

It is critical to secure the immersed probe so that it cannot get tangled up in the agitator.

__________________
They said; "Brain size?" I heard; "Train size?" so I said: "I'll take a small one, thank you."
Register to Reply
Power-User
Engineering Fields - Instrumentation Engineering - New Member Engineering Fields - Control Engineering - New Member Technical Fields - Technical Writing - New Member Technical Fields - Education - New Member Fans of Old Computers - Apple II -

Join Date: Sep 2011
Location: Bellingham, WA
Posts: 191
Good Answers: 46
#22

Re: Liquid Level Measurement in Agitated Tanks

07/18/2013 11:50 AM

Is this a batch-fill process? If so, you can get your second measurement from a flowmeter with an appropriate totalizer. This would also exhibit the advantage of representing liquid level in the same units as the load cell system (i.e. true mass).

__________________
They call me "lightning" when wielding a hammer, because I never strike twice in the same place
Register to Reply
Guru

Join Date: Jun 2009
Location: Bangalore, India
Posts: 725
Good Answers: 24
#24

Re: Liquid Level Measurement in Agitated Tanks

07/19/2013 7:37 AM

If you can have a side tube at the bottom, the agitation will be damped and the level will be steady at the average value.. Measure this level using a suitable device.

__________________
bioramani
Register to Reply
Guru
India - Member - Sensors Technology Popular Science - Cosmology - Dream, Think and Act United Kingdom - Member - New Member United States - Member - New Member Canada - Member - New Member

Join Date: Aug 2006
Location: AM-51, Deen Dayal Nagar, Gwalior, Madhya Pradesh, MP 474001, India
Posts: 3418
Good Answers: 32
#25

Re: Liquid Level Measurement in Agitated Tanks

07/21/2013 11:28 PM

Perhaps a radiation gauge may work as it can sense total material under the path and for averaging you can have larger oppurture and sensing area.

__________________
Prof. (Dr.) Shyam, Managing Director for Sensors Technology Private Limited. Gwalior, MP474001, India.
Register to Reply
Guru

Join Date: Jul 2010
Posts: 669
Good Answers: 176
#28

Re: Liquid Level Measurement in Agitated Tanks

07/22/2013 6:40 PM

>I was actually thinking of having webcams that will take a shot of the level through the calibrated sight glass as a way of validating the load cell reading. So, although the validation is not auto, there is an electronic record of the time when the load cell reading was done and the cam shot.

caveat emptor: A US pharmaceutical advised me that they could not validate bitmap screen shot images of recorded trend data on a CIP cleaning process. The reason - Photoshop - it's possible to alter an image for fraudulent purposes.

I suspect that video files are as subject to editing as static images. You might make inquiries as to what it takes to validate a video file if you pursue that route.

Register to Reply
Associate

Join Date: Mar 2013
Location: North East UK
Posts: 27
Good Answers: 2
#29
In reply to #28

Re: Liquid Level Measurement in Agitated Tanks

07/23/2013 4:03 AM

I agree Validating unsecure records while trying for CFR 21 compliance can be a minefield.

one thing we do onsite for various pieces of equipment is to print and sign results from balances etc so with load cells an auto output to a printer may work. The other route to go down would possibly look to integrating it into a LIMS (Laboratory Information Management System) if the company already uses one as they tend to already be CFR compliant.

Register to Reply
Register to Reply 30 comments

Good Answers:

These comments received enough positive votes to make them "good answers".

"Almost" Good Answers:

Check out these comments that don't yet have enough votes to be "official" good answers and, if you agree with them, vote them!
Copy to Clipboard

Users who posted comments:

Ash55 (8); bioramani (1); dhayanandhan (2); Guest1947 (1); Iris (2); jhhassociates (1); lyn (1); only1yeti (5); Phys (1); PWSlack (4); SHOCKHISCAN (1); Shyam (1); tonykuphaldt (1); Tornado (1)

Previous in Forum: Instrument Ground Philosopy   Next in Forum: Strange Input Values on Eurotherm 3216 Controller

Advertisement