Cable capacitance measurement / calculation

06/25/2008 10:25 AM

I have to verify a cable capacitance spec of a three conductor cable with overall shield measured from all conductors shorted together against the shield.

Unfortunately the wiring test system we use does only perform capacitance measurements between test points (single wire to shield) and does not allow to group test points for this particular test (I guess the system is more intended for actual capacitor measurements).

The idea is to perform three single wire measurements with a new limit value.

The single wire capacitance to shield is 310 pF being the capacitance of all wires shorted to shield 540 pF.

I cannot find a formula for the correlation of the two measurements and I would not like to determine a test limit value based on experiments only.

Can anybody help?

Register to Reply
Pathfinder Tags: cable capacitance
Interested in this topic? By joining CR4 you can "subscribe" to
this discussion and receive notification when new comments are added.
Guru
United Kingdom - Member - Olde Member!! Engineering Fields - Instrumentation Engineering - New Member

Join Date: Jul 2006
Location: Dunstable, England
Posts: 2821
Good Answers: 45
#1

Re: Cable capacitance measurement / calculation

06/25/2008 12:16 PM

Why can't you connect the three cables together and then measure the capacitance to the screen?

__________________
A little knowledge is a dangerous thing - Googling is far worse!
Register to Reply
Active Contributor

Join Date: Jun 2008
Location: Spain / Germany
Posts: 18
#2
In reply to #1

Re: Cable capacitance measurement / calculation

06/26/2008 1:57 AM

I forgot to mention that the cables are terminated with connectors and we are talking about the final test for such assemblies in big quantities. Of course there would always be the possibility to create an adapter to short all three wires for this test but this would require user intervention during an automated test adding to the test time plus the QA control of this adapter.

Register to Reply
Anonymous Poster
#3

Re: Cable capacitance measurement / calculation

06/26/2008 9:19 AM

Capacitance measurements are notoriously inaccurate. Even in a laboratory they are iffy. Capacitor maunfactuters spends much money trying to design test equipment used to test their product and still have problems. Ask yourself what is this data to be used for? Just stuck in some record somewhere and never looked at again? Fugetaboutit!

Register to Reply
Guru
United Kingdom - Member - Olde Member!! Engineering Fields - Instrumentation Engineering - New Member

Join Date: Jul 2006
Location: Dunstable, England
Posts: 2821
Good Answers: 45
#4
In reply to #3

Re: Cable capacitance measurement / calculation

06/26/2008 9:24 AM

I beg your pardon Guest!!

Capacitance measurements are extremely accurate and repeatable if measured correctly!!

I have an old General Radio capacitance bridge that will accurately measure capacitance to within 0.01% absolute value...

And if used to compare capacitance values against a known reference capacitor will measure to within a few parts per million!!!

The total range of measurement is from 10 attoFarads (0.00001pF!!!) to 10 microFarad with resolution to 0.1 ppm

Now say again why you think capacitance measurements are "...a bit iffy..."

John.

__________________
A little knowledge is a dangerous thing - Googling is far worse!
Register to Reply
Active Contributor

Join Date: Jun 2008
Location: Spain / Germany
Posts: 18
#5
In reply to #3

Re: Cable capacitance measurement / calculation

06/26/2008 11:39 AM

Hello Guest,

capacitance measurements of components that have significant parasitic impedances can be indeed a challenge. I have my experiences with varistors and suppressor diodes. High dependance on the measurement voltage, temperature, etc. Also cable capacitance varies with measurement frequency. There are some variables that have to be controlled for repeatable measurements.

Capacitors themselves are built to minimize the parasitic components and maximise stability. If they are built accordingly, they can be measured very accurately as Electroman states.

The capacitance spec of the cables I have to measure is safety related (medical equipment). That´s why I need the records. I know there is a lot of variables that have to be controlled for repeatable measurements.

I will check with some cable manufacturers next to see if they can help.