PH Meters - Calibration and Anomalous Values

09/08/2009 1:38 AM

Hi friends,

I would like to know why similarly calibrated PH meters of different brands read different PH values of a sample but read almost the same value in buffer solutions?

Is there any alternative method, e.g by titration to cross verify the PH and compare it with electronic PH Meter reading? If that method exists, can someone give me details on its procedure?

Anticipated thanks for any help you can provide.

Regards,

Bamboche

Register to Reply
Interested in this topic? By joining CR4 you can "subscribe" to
this discussion and receive notification when new comments are added.

Good Answers:

These comments received enough positive votes to make them "good answers".

"Almost" Good Answers:

Check out these comments that don't yet have enough votes to be "official" good answers and, if you agree with them, vote them!
Guru
Popular Science - Weaponology - New Member United Kingdom - Member - New Member

Join Date: May 2007
Location: Harlow England
Posts: 16512
Good Answers: 670
#1

Re: PH Meter Calibration n Readings Anomaly

09/08/2009 10:27 AM

Good question... ph instruments do seem susceptible to electrical noise, bad cabling and other factors.
Are the samples being read whilst in a system or are they being tested 'on the bench'?
Del

__________________
health warning: These posts may contain traces of nut.
Register to Reply
Member
Popular Science - Weaponology - New Member Engineering Fields - Manufacturing Engineering - New Member

Join Date: Jul 2007
Posts: 5
#3
In reply to #1

Re: PH Meter Calibration n Readings Anomaly

09/09/2009 12:08 AM

Hi Del.

my samples are tested on the bench. I prevail samples of Demin Water for which I am reading different values on my PH meters and worse I cannot say which one is the right one. I have some suspicions about the conductivity of the pure water, which is so close to zero that it may cause the ions migration between the electrodes of the PH Instrument more difficult and which therefore renders the instrument unreliable.

This is probably unfounded theory, or simply can someone enlighten me on that.

Thanks for helping out.

Regards

Bamboche

Register to Reply
Guru
Popular Science - Weaponology - New Member United Kingdom - Member - New Member

Join Date: May 2007
Location: Harlow England
Posts: 16512
Good Answers: 670
#4
In reply to #3

Re: PH Meter Calibration n Readings Anomaly

09/09/2009 3:20 AM

I'm not that au fait with ph meters, but I suspect that the various factors like ph and conductivity do interact.
I guess the answer is to 'manage' the situation by standardizing on one meter, and maybe relegate the other meters for less critical duties.

Dunno if it's of interest but we were using conductivity to measure concentration of a vegetable washing solution which was mostly ascorbic and citric acid. The conductivity went up nicely with concentration.... And then started coming down again as concentration increased.
Del

__________________
health warning: These posts may contain traces of nut.
Register to Reply
Guru
United Kingdom - Member - Indeterminate Engineering Fields - Control Engineering - New Member

Join Date: Jan 2007
Location: In the bothy, 7 chains down the line from Dodman's Lane level crossing, in the nation formerly known as Great Britain. Kettle's on.
Posts: 32175
Good Answers: 839
#12
In reply to #3

Re: PH Meter Calibration n Readings Anomaly

09/11/2009 11:50 AM

Ah! The pH of demin water is not particularly relevant. It's the same as asking, "what is the concentration of the hydrogen ion concentration in a solution that has so little of anything else in it that the only substances present are hydrogen ions, hydroxide ions and water molecules?"

If there were some acid-making substances present, then the water would not be particularly demineralised. So use the conductivity instrument instead!

__________________
"Did you get my e-mail?" - "The biggest problem in communication is the illusion that it has taken place" - George Bernard Shaw, 1856
Register to Reply
Anonymous Poster
#2

Re: PH Meters - Calibration and Anomalous Values

09/08/2009 11:53 PM

This seems crazy, but this occured with me as well. Temperature and product density gave two different results, but the buffer solution was exactly the same reading. By sampling several samples the electronic pH tester self calibrated and produced the same result. Hope this helps!

Register to Reply
Anonymous Poster
#5

Re: PH Meters - Calibration and Anomalous Values

09/09/2009 8:51 AM

You need to perform a minimum of two point calibration with two buffers, usually 7 and 4 or 7 and 10. Try to bracket your expected sample value with the calibration buffers. The pH meter should then report an efficiency, usually as a percentage slope. Anywhere from 94 to 101 % is acceptable. Also use the pH 7 buffer first to set the zero potential and then the 4 or 10 to set the electrode slope. If both meters are calibrated the same way and both have acceptable slopes they will measure the unknown pH very closely.

Register to Reply
Active Contributor

Join Date: Apr 2009
Posts: 14
Good Answers: 2
#6

Re: PH Meters - Calibration and Anomalous Values

09/09/2009 10:30 AM

The ions in solutions produce a potential across the electrodes. Buffer solutions are usually simple small ions but samples sometimes have ions, or other molecules, dispersed or dissolved in them which interfere with ion transport at the electrodes, and as a result change the electrical signal from the electrodes. The meter is actually measuring a tiny electrical potential across the electrode membranes. Since different electrodes will make slightly different electrical potentials in various sample solutions, it is possible to get variations in "measured" pH. Even electrodes which are supposed to be the same and made of the same materials can cause this if the electrodes have a different history. Various solutions can plug electrode tips or contaminate them with something that does not wash off easily. Soaking electrodes for a day or two in water or dilute HCl can sometimes help. The differences in measured pH should however not be very large (more than a tenth or two of a pH unit), and if they are, then something is hinkey (scientific term for messed up).

Register to Reply
3
Power-User

Join Date: Dec 2006
Posts: 442
Good Answers: 32
#7

Re: PH Meters - Calibration and Anomalous Values

09/09/2009 11:30 AM

It is virtually impossible to measure the pH of deionized water using a pH meter. Other methods such as titration and measuring resistance may give more accurate results if done very carefully, but the results still would have little meaning in a practical sense.

First, the measuring problem: A pH meter is a voltmeter with extremely high internal resistance. For a voltmeter to be accurate, the resistance across which the voltage is being measured must be significantly lower than the internal resistance of the voltmeter. Otherwise, the voltmeter itself acts as a parallel resister, and an inaccurately lower voltage is indicated by the voltmeter.

With deionized water, the resistance of the probe/ DI water is close to the internal resistance of the meter, so the reading will have error. Meters with higher internal resistance will give more accurate readings. However, I would suspect that even the same meter using different probes will give different readings, due to variations in probe resistance. This is not a problem with buffers, which have high ionic strength (low resistance).

DI water has a pH of 7. Anything that is introduced to the water will cause significant changes in pH, because it has no buffering capacity. Exposing the water to air causes it to absorb CO2, which lowers the pH. Putting the water in a beaker changes the pH, as small amounts of the container dissolve/leach ions into the water. Even the act of putting pH probes into the water changes the pH, as the reference electrode releases ions into the water.

If you are making your measurements on the bench, it is certain that the pH you measure is not the same as the pH in the bulk container of water, even if you could make a perfect pH measurement. So the reading is pretty much meaningless in the real world.

Here are some tips: link

Adding a small amount of absolutely pure KCl to the water should increase the ionic strength (and therefore accuracy and stability) of the reading without changing pH. But, in the end, what does the reading really tell you?

Tad

Register to Reply Good Answer (Score 3)
Member
Popular Science - Weaponology - New Member Engineering Fields - Manufacturing Engineering - New Member

Join Date: Jul 2007
Posts: 5
#9
In reply to #7

Re: PH Meters - Calibration and Anomalous Values

09/10/2009 1:51 AM

Hi Tad,

I thank you very much for the enlighting information that you have given. There is therefore a closely related link between Conductivity and PH-Measurements, since it involves the use of a sensitive voltmeter in a media that contains almost only H+ and OH- ions together with traces of contaminants that will alter the PH values. So far, so good; If I take your point, it will mean that when the conductivity of the Pure water is very low, that is, with low ions migration activity, the PH-Meter will be less effective. This is characterised by a longer stabilisation time to read PH. And I suppose that every PH meter from different suppliers will give different results (as it is actually the case at my place).

My questions:

1) Can you suggest alternative methods for measuring PH of Pure Water, such as your titration method. If the method can provide a means to determine the error values that are related to the drift of the PH-Meters.

2) Is there any type of Resistance that can be coupled with the probe to increase the internal Resistance?

3) To conclude, I still have doubts about the End result, I am not sure what will be measured?

Register to Reply
Power-User

Join Date: Dec 2006
Posts: 442
Good Answers: 32
#10
In reply to #9

Re: PH Meters - Calibration and Anomalous Values

09/10/2009 10:31 AM

Unfortunately, I do not have a good answer for you. I would think there is a colorimetric method that might work, but I could not find any specific references for ultrapure water.

I noticed that some of the pH probe manufacturers market pH probes designed specifically for this application. Try Googling pH and deionized water or ultrapure water.

It appears that ASTM, USP and other organizations have methods and procedures for making these measurements. This paper goes into some of the problems, and provides references for some of the standard methods.

I would get copies of the USP and ASTM methods, and use one of these procedures. If your results are ever questioned, it is hard to argue about results when using approved standardized procedures.

I personally feel that you are being asked to make a measurement which really doesn't make much sense, or provide any useful information. Please note that ISO has no requirement to measure pH of ultrapure water for this reason.

However, I know specs are specs, and it is seldom worth arguing over their validity. The best solution in this case is to use the ASTM or USP method. At least your butt is covered by using a "validated" standard procedure.

Good luck, and let us know how you make out.

Tad

Register to Reply Score 1 for Good Answer
Member

Join Date: Aug 2009
Posts: 7
Good Answers: 1
#8

Re: PH Meters - Calibration and Anomalous Values

09/09/2009 4:45 PM

Different electrodes will give different responses. Good quality liquid KCl electrodes will be more accurate than gel electrodes. Temperature compensation algorithms and temperature sensor quality will also impact on the measurement of solutions, whereas buffers are normally measured at 25 deg C.

There is no alternative measurement technique for pH. Your best bet is a good quality instrument with the best electrode and buffers you can afford.

Radiometer of Copenhagen make the best instruments and their IUPAC buffers are the closest you'll come to a NIST standard.

Register to Reply
Member

Join Date: Aug 2009
Posts: 7
Good Answers: 1
#11
In reply to #8

Re: PH Meters - Calibration and Anomalous Values

09/10/2009 3:22 PM

It is impossible to measure the pH of deionised water using a bench application. The DI water will immediately start to absorb CO2, which will depress the pH. Furthermore, any contaminants in the sample container will tend to equilibriate with the solution, thus influencing the pH. Adding KCl is NOT recommended, unless you are analysing the pH of soils or other normally solid substrates. The only way to accurately determine the pH of DI waters is to use a good quality on line analyser. The best analysers are made by companies such as Swan and Polymetron. The best alternative method uses a dual conductivity ( specific and cation conductivity ) analyser and calculates the pH using an algorithm specific to the solution. This is not normally a recommended technique for DI waters, because of the relatively small difference between SC & CC, but they can still be of use.

There are no titrimetric or colourimetric techniques to accurately determine pH, so don't even waste your time with them.

Register to Reply
Register to Reply 12 comments

Good Answers:

These comments received enough positive votes to make them "good answers".

"Almost" Good Answers:

Check out these comments that don't yet have enough votes to be "official" good answers and, if you agree with them, vote them!
Copy to Clipboard

Users who posted comments:

Anonymous Poster (2); bamboche (2); marcus43 (2); PWSlack (1); Robert Trout (1); Tad (2); user-deleted-1105 (2)

Previous in Forum: PLC Problem   Next in Forum: Seal Pot Connection with High-Pressure Side of Instrument
You might be interested in: pH Electrodes, pH Transmitters, pH Instruments

Advertisement